US Recall:

Announcements of Enforcement by the Food and Drug Administration pertain to action taken (i.e. Market withdrawal) in connection with agency regulatory activities.

If you have been injured by the Calaxo interference screw implant or wish to inquire about individual actions directed against
Smith & Nephew regarding said implant please use our Case evaluation form or Failed Medical Device formto request further information.

US RECALL

MANUFACTURER: Smith & Nephew, Inc., Mansfield, Massachusetts.


PRODUCT: Acufex Tibial anchor Screw & Spiked Washer, indicated for use in fixation of soft tissue in orthopedic procedures. Recall #Z-1011-9.


REASON: The head of the screw may break during insertion.


***

Medical Device/Equipment ALERT



Past Notices:



Ref. MDEA(NI)2007/82
Issued: 4th October 2007
For: IMMEDIATE ACTION 􀀹
ACTION
UPDATE
INFORMATION
Section
Medical Device/Equipment:
Bioresorbable screw: CALAXO interference screw implants
manufactured by Smith & Nephew Inc.

Problem:
Recall due to pre-tibial soft tissue swelling after implantation of CALAXO screws leading to the
need for surgical intervention to remove any remaining screw fragments.

Action by:
Orthopaedic surgeons
Supplies managers
Theatre managers

Action:
Identify and quarantine any unimplanted affected devices.
Do not implant affected devices.
Return affected devices to the manufacturer. Do not implant affected devices.

Distributed by NIAIC to:
Chief Executive of each HSS Board
Chief Executive of each HSS Trust
Chief Executive of each Agency
NIAIC Liaison Officers

For onward distribution see Section 5
Contacts
Details of manufacturer contacts and NIAIC contacts for technical aspects
Feedback Requirements to NIAIC
In accordance with PEL(06)17 acknowledgment of assurance should be given:-
This Alert is on our web site: http://www.dhsspsni.gov.uk/niaic
Page 2 of 3 pages
1. DEVICE/EQUIPMENT:
Recall due to pre-tibial soft tissue swelling after implantation of CALAXO screws leading to the
need for surgical intervention to remove any remaining screw fragments.
All lot numbers and all item codes distributed since 2006.
2. PROBLEM:
Smith & Nephew Inc. is recalling all lot numbers and all item codes of CALAXO screws which were
first distributed in the UK in 2006. This screw is used to secure the graft in anterior cruciate
ligament (ACL) reconstruction.
Worldwide, 0.3% of patients have developed pre-tibial soft tissue swelling, which can mimic the
appearance of an infection between 2 and 36 weeks after implantation. The manufacturer is
currently investigating the cause of this problem.
Most cases resolved without intervention. A small number of cases needed local debridement and
removal of any remaining screw fragments, or replacement with an alternative screw or bone graft.
The manufacturer estimates that about 2,550 devices have been used in the UK since 2006.
The manufacturer is aware of seven reported cases of inflammation in the UK of which two
required implant revision.
The manufacturer wrote to chief executive officers of trusts, theatre managers, and surgeons who
were known to use CALAXO about this product recall in August 2007 (see MHRA website). This
notice is to facilitate the recall.
3. ACTION BY:
Orthopaedic surgeons
Supplies managers
Theatre managers
4. ACTION:
Identify and quarantine any unimplanted affected devices.
Do not implant affected devices.
Return affected devices to the manufacturer.
Identify all patients implanted with this screw.
If they present with pre-tibial swelling within 2-36 weeks of implantation, consider:
􀂃 aspirating the area for routine cultures
􀂃 local debridement and removal of any remaining screw fragments
􀂃 replacement with an alternative screw or bone graft if necessary.
Report all adverse incidents related to CALAXO screws to the implant manufacturer and the
NIAIC.
5. ONWARD DISTRIBUTION TO:
Please bring this notice to the attention of all who need to know or be aware of it. This will include
distribution to:
Infection prevention and control directors
Medical directors
Orthopaedic surgeons
Risk managers
Supplies managers
Theatre managers
Theatres
Independent Health and Social Care
Providers – Private Hospital & Clinics
through RQIA
Page 3 of 3 pages
6. CONTACTS:
Enquiries to the manufacturer should be addressed to:
Dr Mike Eyre
Smith & Nephew Healthcare Limited
Healthcare House
Goulton Street
Hull HU3 4DJ
Tel: 01482 222 200
Fax: 01482 222 211
E-mail:
Mike.Eyre@smith-nephew.com
Enquiries to NIAIC should quote reference number MDEA(NI)2007/82 and be addressed to:
Northern Ireland Adverse Incident Centre
(NIAIC)
Health Estates
Estate Policy Directorate
Stoney Road
Dundonald
Belfast BT16 1US
Tel: 028 9052 3868
Fax: 028 9052 3900
Email:
NIAIC@dhsspsni.gov.uk
7. FEEDBACK:
In accordance with PEL(06)17 the following acknowledgment of assurance should be given:-
Deadline (Email received) : 8th Oct 2007
Deadline (action underway) : 18
th Oct 2007
Deadline (action complete) : 4
th Dec 2007
Robert Sergeant
NIAIC Operational Manager
HOW TO REPORT ADVERSE INCIDENTS
Adverse Incidents relating to medical devices, non-medical equipment, plant and buildings should be reported to NIAIC as soon as possible.
Advice on how to report is given in MDEA(NI)2006/01. If you are in doubt about how to report incidents, please speak to your liaison officer or
contact NIAIC using the telephone number provided. Adverse Incident reporting forms and an on-line reporting facility are available on the
NIAIC website at www.dhsspsni.gov.uk/niaic
Heath Estates is an Executive Agency of the Department of Health, Social Services and Public Safety

See the original: http://www.dhsspsni.gov.uk/mdea_ni_2007-82.pdf